any ideas where to look for filters ? theyre just a flat disc maybe 45mm dia or so
snap on says no
snap on says no
Are these the ones you need?
https://www.ebay.co.uk/itm/Bluepoin...800363?hash=item442d1f45eb:g:Rk4AAOSwyRdb6f~V
hi there i have just found this post i have a mca1200 gas analyser and its in service mode is there anyway this can be switched off I'm not an mot garage any info would be great have tried snap on and searched to no a vale cheers DaveI’ve not worked on one of these particular analysers, but used to calibrate 4 gas analysers for MOT stations. Warm up time on older analysers could be anything up to 30 minutes. During this time, the readings can fluctuate all over the place. If the readings are not stable after the warm up period, try running it (fresh air only)without the paper filters in as they can get impregnated with hydro-carbons. If the readings are still unstable, then it’s likely that the infra red gas bench has been flooded at some point. On some makes, it is possible to salvage a previously flooded bench by stripping it down and carefully cleaning out the gas sample tube and the i/r source and filters.